Hi Jingfu, pulsed measurements in qudi-iqo-modules allow to use the sequence mode of an AWG. With this single "waveforms" are created (termed For T1 measurements in particular, note that there are three different generation methods in the basic_predefined_methods of the qudi-iqo-modules. One creating an ensemble with equally spaced points (line 637 in the file) and one creating an ensemble but with logarithmically spaced points (line 697). Logarithmic spacing has the advantage that more points for shorter times are measured where more absolute decay is expected. The third one (line 1003) writes a PulseSequence with logarithmically spaced points. Here AWG memory is kept very low by using a single idle-ensemble that is repeated integer k times for different data points where k is exponentially growing. It is possible to create any own, custom |
Beta Was this translation helpful? Give feedback.
-
Please note that sequence mode works well for your issue as long as you do not care about phases between sequence steps (e.g. MW pulses). If you spread MW pulses over multiple steps, jitter from the AWG sequencer becomes a major concern. Depending on your device specs, this might or might not work. |
Beta Was this translation helpful? Give feedback.
-
Thanks. |
Beta Was this translation helpful? Give feedback.
-
thanks Ilia. |
Beta Was this translation helpful? Give feedback.
-
In the current version for Qudi, we generate the whole pulse sequnece in one single waveform, e.g., for 100 points in T1 measurement. There are surely some advantages. However, there are also problems from the limitation of AWG memory size. Could we generate one waveform for one point, i..e, for one delay time? In this way, we can measure much more points for one curve.
Thanks.